Bioelectrical impedance vector analysis in critically ill patients: a prospective, clinician-blinded investigation

Abstract

Introduction

Assessment of fluid status in critically ill patients is challenging. We aimed to assess the feasibility and validity of bioelectrical impedance vector analysis (BIVA) as a measure of hydration in critically ill patients.

Methods

We performed twice-daily BIVA measurements and fluid balance calculations and recorded physiological variables in mechanically ventilated patients within 24 h of intensive care unit (ICU) admission for up to 5 days. Treating clinicians were blinded to BIVA results.

Results

We performed 344 BIVA measurements in 61 patients. According to BIVA, 14 patients (23 %) were dehydrated, 22 (36 %) were normally hydrated and 25 (41 %) were overhydrated upon ICU admission. Patients with normal BIVA hydration were less sick, had fewer comorbidities and had less deranged physiology than patients found to be dehydrated or overhydrated with BIVA. Cumulative fluid balance increased in patients found to be dehydrated with BIVA by a mean of 3.4±2.2 L, whereas in patients found to be overhydrated with BIVA, it decreased by a mean of 4.5±6.9 L. In patients found to be normally hydrated with BIVA, fluid balance remained unchanged. BIVA-defined hydration increased with 1 L (median change 1.5 %, P =0.09) or 2 L (median change 0.7 %, P =0.09) of calculated fluid gains. BIVA-defined hydration decreased (median change −0.8 %, P =0.02) with a negative cumulative fluid balance of >2 L. BIVA-defined hydration between first and last measurement correlated with the corresponding change in fluid balance (ρ =0.25, P =0.05).

Conclusions

BIVA is feasible in critically ill patients. Its validity is supported by the observed characteristics of patients with different degrees of BIVA hydration upon admission and by different fluid management of such patients by blinded clinicians. The sensitivity of repeated BIVA hydration measurements to detect fluid accumulation or fluid balance changes <2 L was low, however. These contradictory findings provide the rational basis for studies of BIVA-assisted fluid management in ICU patients.

Introduction

Fluid management in patients who are critically ill is challenging because their hydration status is difficult to assess at the bedside. Widely used markers such as invasively obtained intravascular pressures are acknowledged to have major flaws as measures of hydration [13], and they provide no information about extravascular or intracellular fluid status. Furthermore, calculated cumulative fluid balance during intensive care unit (ICU) admission may be inaccurate, does not account for insensible or third-space fluid losses and does not incorporate preadmission fluid status. Standardised body weight measurement, even using beds with electronic weight estimation, have repeatedly been shown to be divorced from cumulative fluid balance calculations [46]. Clinical detection of fluid overload as oedema is insensitive and requires the accumulation of 4–5 L before detection [7]. Finally, deuterium dilution studies, which are the gold standard for total body water (TBW) assessment, are not feasible on a daily basis in the ICU setting [8].

Bioelectrical impedance vector analysis (BIVA) is a rapid, non-invasive bedside technique to measure fat-free TBW that correlates closely with the deuterium dilution technique [9, 10]. BIVA has been reported to be useful to monitor hydration status during fluid removal in patients with decompensated heart failure [11, 12] and during intermittent haemodialysis [1317]. Thus, it may also prove useful in patients who are critically ill.

However, there is a paucity of studies of the feasibility and validity of BIVA-derived measurements in critically ill patients, and no study to date has examined the association between BIVA-measured hydration status and changes in fluid balance or clinician-driven fluid management. Accordingly, we conducted a prospective, clinician-blinded, observational study to assess the feasibility and validity of BIVA in critically ill patients.

Materials and methods

Study design

We conducted a prospective, clinician-blinded, observational study from August 2014 to February 2015 to evaluate BIVA-measured hydration status in adult ICU patients at a tertiary hospital. The study was approved by the Austin Hospital Human Research Ethics Committee (reference number H2012/04864). Informed consent was obtained from the person legally responsible for the patient.

Inclusion and exclusion criteria

We enrolled a consecutive convenience sample of screened patients by eligibility on weekdays only. We included adult patients (≥18 years) within 24 h of admission to the ICU who were mechanically ventilated and were expected to stay in the ICU for ≥48 h.

We excluded patients with end-stage kidney disease undergoing long-term dialysis, pregnant patients, patients admitted to the ICU following elective cardiac surgery, patients with pacemakers or implantable defibrillators and patients with diaphragmatic pacing.

Study protocol

We performed twice-daily (morning and afternoon) BIVA measurements using Renal EFG BIVA™ Technology (EFG Diagnostic, Belfast, UK) during the first 5 days in the ICU or until ICU discharge. At the time of each BIVA measurement, we recorded mean arterial pressure (MAP), central venous pressure (CVP), vasopressor requirements, gas exchange, arterial lactate, creatinine and cumulative fluid balance.

Three trained operators obtained the BIVA measurements according to the manufacturer’s instructions with patients positioned horizontally and supine whenever possible for ≥2 minutes. BIVA results were not made available to treating clinicians at any time during the study.

Principles of Renal EFG with BIVA™ Technology

BIVA provides a quantitative estimate of TBW as a percentage of fat-free body mass. It combines bioelectrical impedance with capacitance measures (i.e., time required to charge a circuit). Whole-body impedance is considered as a combination of resistance (R) and reactance (Xc). The arc tangent of Xc/R is called the phase angle and represents the phase difference between voltage and current [18].

BIVA is safe and non-invasive and can be performed at the bedside, giving a result within minutes. It is based on the electrical principle that the body is a circuit with a given resistance (opposition of current flow through intracellular and extracellular solution) and reactance (capacitance of the cells to store energy) [19]. Through the application of a 50-kHz current to the body via a pair of electrodes (one placed on the dorsum of the wrist and the other on the ipsilateral ankle), it gives a measurement of TBW in fat-free tissues.

The accuracy of the test rests on careful placement of the BIVA electrodes and the connecting cable clips. Standardised by height (H), BIVA results are displayed graphically, comparing R/H to Xc/H and reflecting hydration abnormalities and soft tissue mass. If obtained as a single measurement, the BIVA result can be compared with the normal population, represented on the graph by confidence ellipses, with data expected to fall within the reference 75 % tolerance ellipse. Relative hydration can also be depicted as vector length. Longer vectors equate to volume depletion, whilst shorter vectors correspond to volume overload [19].

In our study, patients were classified by BIVA as dehydrated (<72.7 % of fat-free body mass), normally hydrated (hydration 72.7–74.3 %) or overhydrated (hydration >74.3 %). Examples of values obtained with the Renal EFG BIVA™ Technology are displayed in Fig. 1.

Fig. 1
figure 1

Examples of values obtained with the Renal EFG BIVA™ Technology. BIVA bioelectrical impedance vector analysis, H height, PA phase angle, Xc reactance, BCM body cell mass

Statistical analysis

Statistical analyses were performed using STATA version 11.2 software (StataCorp, College Station, TX, USA). Continuous variables were summarised using median and interquartile range (IQR) or mean ± standard deviation, and categorical variables were expressed as number (%). The Kruskal-Wallis test was used for comparison between continuous variables, and the χ2 test or Fisher’s exact test was used for comparisons between categorical variables. Changes over time for multiple measures of cumulative fluid balance were tested by repeated-measures analysis of variance (RM-ANOVA) using ICU day as the RM variable. For comparison of cumulative fluid balance change over time between categories, an interaction variable (between category and time) was introduced in the RM-ANOVA model. The nonparametric Wilcoxon signed-rank test was used to assess the change between two prespecified time points. Spearman’s rank correlation coefficient was used to assess the relationship between changes in BIVA hydration, cumulative fluid balance, CVP and lactate. Intra-BIVA coefficient of variation between duplicate measurements obtained under steady-state conditions was calculated as described in Additional file 1. A P value <0.05 was considered statistically significant.

Results

Patient characteristics

We screened 1292 patients and identified 73 mechanically ventilated patients admitted to the ICU for <24 h (Fig. 2). The restrictive inclusion criteria explain the relatively low screening/inclusion ratio. Firstly, we did not include 231 patients admitted after cardiac surgery between August 2014 and February 2015. Secondly, approximately 50 % of patients admitted to the unit did not require mechanical ventilation and were therefore not eligible. Thirdly, we did not include patients admitted on weekends or during holidays. Finally, some patients were subject to multiple invasive procedures and investigations and therefore could not be included within 24 h of ICU admission. After excluding 8 patients who declined participation and 4 patients in whom only one BIVA measurement could be performed owing to death or early discharge from the ICU, 61 patients were included in the final analysis. We included these patients within a median of 14 (IQR 7 to 19) h and obtained a total of 344 BIVA measurements. Their median age was 63 years and the median Acute Physiology and Chronic Health Evaluation III score was 64 (Table 1).

Fig. 2
figure 2

Selection of study patients. BIVA bioelectrical impedance vector analysis, ICU intensive care unit

Table 1 Baseline characteristics of the study cohort, stratified by hydration level according to bioelectrical impedance vector analysis measurement upon admission

At the time of inclusion, BIVA identified 14 patients (23 %) as dehydrated, 22 (36 %) as normally hydrated and 25 (41 %) as overhydrated. The characteristics of study patients according to their level of BIVA hydration at inclusion are summarised in Table 1. Asthma and chronic obstructive pulmonary disease were more common among dehydrated patients than in the other patients (P =0.007). In contrast, chronic liver disease was more frequently observed in overhydrated patients than among the other patients (P =0.009). Similar findings applied to patients admitted with non-operative respiratory and gastrointestinal diagnoses.

Physiological state on admission

BIVA values, biochemistry, haemodynamics and ventilatory variables at inclusion are summarised in Table 2. Upon admission, overhydrated patients required more noradrenaline than dehydrated and normally hydrated patients (P =0.02). Moreover, kidney dysfunction was more pronounced in overhydrated patients than in the other patients (P =0.03). Three dehydrated patients and nine normally hydrated patients were enrolled in the study before a central venous catheter was inserted. Admission CVP was therefore missing in these patients.

Table 2 Admission bioelectrical impedance vector analysis data, haemodynamics, oxygen exchange and biochemical data

Changes in fluid balance and BIVA hydration

Changes in cumulative fluid balance according to BIVA hydration are displayed in Fig. 3. By day 4, blinded clinician driven cumulative fluid balance had increased in patients found to be dehydrated with BIVA (P =0.01 by RM-ANOVA) by a mean of 3.4±2.2 L. In contrast, the cumulative fluid balance in patients found to be overhydrated with BIVA had decreased by a mean of 4.5±6.9 L (P <0.001 by RM-ANOVA). In normally hydrated patients, the fluid balance remained unchanged. The components of the cumulative fluid balance (fluid input and output), urine output and loop-diuretic administration are displayed in the Additional file 1.

Fig. 3
figure 3

Cumulative fluid balance during the first 4 days in the intensive care unit (ICU) in patients with dehydration, normal hydration and overhydration on admission. Values are mean ± SE. BIVA bioelectrical impedance vector analysis

A trend toward increased BIVA hydration over time was observed in patients classified as dehydrated upon admission (P =0.45). In contrast, BIVA hydration significantly decreased in patients with BIVA overhydration upon admission (Fig. 4).

Fig. 4
figure 4

Mean daily hydration determined by bioelectrical impedance vector analysis (BIVA) in patients with dehydration (a), normal hydration (b) and overhydration (c) upon admission. P values represent repeated-measures analysis of variance for change in hydration level over time

Relationship between changes in fluid balance and BIVA hydration

Of the 61 patients, 26 (43 %) reached a cumulative fluid balance of at least +1000 ml (median 1385 ml, IQR 1205–2022 ml) within a median of 1 day after the first BIVA measurement (Table 3). Simultaneously, median BIVA hydration also increased from 73.8 % to 79.7 % (P =0.09). During the same observation period, there was no significant change in CVP or lactate. In 13 (21 %) of 61 patients, cumulative fluid balance reached at least +2000 ml (median 2419 ml, IQR 2196–2696 ml) within the same time frame. The corresponding BIVA hydration also increased from 73.1 % to 73.8 % (P =0.09). However, no significant changes in CVP or lactate were observed. In the 19 patients (31 %) with a negative fluid balance of −1000 ml, a small decrease in BIVA hydration was seen (P =0.22). In patients with a negative fluid balance of at least −2000 ml, we observed a more pronounced decrease in BIVA hydration, from 77.3 % to 73.9 % (P =0.02). There were no significant corresponding directional changes in CVP or lactate.

Table 3 Changes in hydration determined by bioelectrical impedance vector analysis in patients with different magnitude changes in cumulative fluid balance

We also considered the changes in BIVA hydration, CVP, lactate and cumulative fluid balance between the time points of the first and last BIVA recordings (Fig. 5). We found a significant but weak correlation between changes in fluid balance and changes in BIVA hydration, as well as changes in lactate, but no correlation with changes in CVP. In addition, we found no significant correlation between changes in BIVA hydration and changes in lactate.

Fig. 5
figure 5

Changes in hydration determined by bioelectrical impedance vector analysis (BIVA) (a), central venous pressure (CVP) (b) and lactate (c) versus changes in cumulative fluid balance and changes in BIVA hydration versus changes in lactate (d) between the time points of the first and last BIVA recordings

Complete longitudinal data on BIVA hydration and cumulative fluid balance was available for two patients treated with continuous renal replacement therapy (CRRT; one of whom was treated simultaneously with extracorporeal membrane oxygenation). These patients’ BIVA and fluid balance trajectories are displayed in Additional file 1.

Association of BIVA hydration and patient outcomes

Patient outcomes are shown in Table 4. A non-significantly lower proportion of normally hydrated patients received CRRT (P =0.80), died in the ICU (P =0.89) and in hospital (P =0.55) than dehydrated and overhydrated patients.

Table 4 Outcome data

Discussion

Key findings

We conducted a pilot, observational, clinician-blinded study of BIVA-measured hydration status in critically ill patients. We found that, on initial BIVA assessments, 40 % of patients were overhydrated and almost one-fourth were dehydrated. Patients with abnormal BIVA values had specific features congruent with BIVA measured status and appeared more severely ill than patients with normal BIVA status. Importantly, their subsequent mean cumulative fluid balance, as achieved by clinicians blinded to BIVA results, was positive for dehydrated patients, negative for overhydrated patients and neutral for normally hydrated patients. Moreover directional changes in BIVA were consistent with directional changes in fluid balance. Finally, BIVA correlated with cumulative fluid balance over the time in the ICU.

Relationship to previous studies

To our knowledge, hydration status has been measured in ICU patients using BIVA technology in only three other studies [2022]. Basso et al. reported a 70 % incidence of overhydration among 64 critically ill patients [21]. However, surprisingly, the hydration status of these patients remained largely unchanged during the observation period. Unfortunately, no separate information about dehydrated and normally hydrated patients was provided, and no data were presented on cumulative fluid balance in the different groups, to provide information on the validity of BIVA measurements.

In a previous study of 121 patients published in 2000 using older technology [20], single CVP and BIVA measurements were determined simultaneously. CVP values were significantly and inversely correlated with individual impedance vector components and with both vector components together. However, in this study, no additional BIVA measurements were obtained, and no information on changes in overall hydration status or fluid balance was provided.

In a recent study, fluid status was assessed using BIVA before the commencement of CRRT in 58 critically ill patients and 3 days afterward [22]. Measurements of serum N-terminal pro-B-type natriuretic peptide (NT-pro-BNP) were taken at the same time points. Patients were categorised as having either no overhydration or overhydration, but the specific BIVA cutoff values were not given and no separate information about dehydrated patients was provided. Similarly to our study, 50 % of patients were overhydrated before the start of CRRT according to the BIVA measurement and also had an abnormal NT-pro-BNP measurement. Cumulative fluid balance over 3 days was measured and was positive in all groups, but serial BIVA measurements over the same time frame were not obtained. Finally, in all the above studies, treating clinicians were not formally kept blinded to BIVA findings, creating an important risk of bias.

A relationship between fat-free TBW gain or loss and corresponding changes in BIVA hydration has been described in non-ICU patients. A 6.8-L TBW gain during pregnancy, quantified by deuterium dilution, was associated with a significant increase in BIVA hydration [10]. The sensitivity of BIVA to detect smaller hydration changes is uncertain, however. In the present study, we demonstrated an increase in BIVA hydration in patients with calculated fluid accumulations >1 L. Moreover, we observed a statistically significant decrease in BIVA hydration after a calculated median fluid loss of 2.4 L. This supports the results of a study by Piccoli and coworkers [17], who observed decreased hydration (increased vector length) in parallel with fluid removal of 2.4 L during haemodialysis.

Implications of study findings

We have demonstrated the feasibility of performing repeated BIVA measurements in critically ill patients. Additionally, we have demonstrated that an abnormal BIVA value separates logical patient groups according to expected hydration status, that directional changes in BIVA hydration are consistent with directional changes in fluid balance, that changes in BIVA hydration are coherent with changes in fluid balance, and, finally that a change in BIVA over the duration of the ICU stay correlates with cumulative fluid balance. These observations support the face validity, concurrent validity, construct validity and content validity of BIVA as a measure of hydration in critically ill patients.

Our results suggest that BIVA may add useful information to guide fluid management in critically ill patients. However, although the directional changes in BIVA hydration corresponded with directional changes in cumulative fluid balance, a fluid loss >2 L was required for a corresponding decrease in BIVA hydration to reach statistical significance. This may imply that BIVA is insensitive to smaller fluid balance changes or, alternatively, reflects the imperfection of fluid balance calculations. Whether BIVA-guided fluid management improves patient-centred outcomes needs to be explored in future interventional studies.

Finally, we found a significant positive correlation between changes in lactate and cumulative fluid balance over the entire study period. Rather than being related to fluid status per se, this association is suggested to be related to illness severity because both increasing lactate and progressive fluid overload are typically observed in sicker patients. In fact, as demonstrated in Table 3, no significant changes in lactate during early fluid gains or losses >1–2 L were found. Moreover, we found no correlation between changes in lactate and changes in BIVA hydration.

Strengths and limitations

Our study has several strengths. By obtaining twice-daily measurements over several consecutive days, we were able to exactly match changes in fluid balance with corresponding changes in BIVA hydration, CVP and lactate and thereby study their relationships. Since treating clinicians were blinded to the results, it was possible to do an unbiased comparison of patients’ characteristics and outcomes at different BIVA hydration levels. The observation that patients with dehydration, normal hydration and overhydration on initial BIVA assessment had different characteristics that could be expected owing to such differences in hydration status further supported our findings. Moreover, the fact that patients with dehydration, normal hydration and overhydration on initial BIVA assessment had a subsequent concordant change (positive, neutral or negative) in cumulative fluid balance supports the validity of BIVA. Finally, the observation that directional changes in BIVA hydration corresponded with directional changes in fluid balance lends robustness to our study.

This study has limitations. It is a single-centre study, with all the limitations inherent in such a design. However, we studied a heterogeneous population of ICU patients with >300 measurements, suggesting a degree of external validity and robustness. It is not an interventional study; therefore, we can make no inferences about the utility of BIVA in the management of fluid balance. However, observational studies such as this one are necessary to establish the feasibility, safety and validity of the technique before its application in interventional studies. We did not use the deuterium dilution method to confirm the accuracy of the fat-free TBW assessments provided by the BIVA technology. However, the lack of steady state in critically ill patients and the time taken for equilibration preclude the simultaneous use of gold standard isotopic tracers as a formal validating technique. In addition, the significant number of missing CVP values upon admission in dehydrated and normally hydrated patients needs to be acknowledged. This may have biased the relationships of CVP with BIVA hydration and cumulative fluid balance, respectively.

Whilst the BIVA measurements of hydration status mirror the changes seen in cumulative fluid balance, we acknowledge that the ICU environment itself may impact the accuracy of BIVA measurements. In some circumstances, patients could not be positioned completely supine (e.g., those with head injury and intracranial pressure monitoring), and on occasion the positioning of the electrodes had to be modified slightly owing to the presence of other devices (e.g., intravenous cannulae and soft restraints). The extensive electrical equipment in the ICU, including the various monitoring devices and mechanical ventilators, could potentially impact measured bioimpedance, as could any water in the patient’s bed, though to what degree is unknown. The fact that in this technically hostile environment BIVA still generated reproducible and logical findings supports the applicability of this technology in the ICU.

Conclusions

BIVA hydration may be an additional measure of fluid status in critically ill patients. Two important findings support this conclusion. Firstly, patients with dehydration, normal hydration and overhydration at initial BIVA assessment had different characteristics that could be expected from such differences in hydration status. Secondly, patients with dehydration, normal hydration and overhydration in their initial BIVA assessment had subsequent changes (positive, neutral or negative) in cumulative fluid balance that were consistent with clinical expectations. However, our findings challenge the sensitivity of repeated BIVA hydration measurements as a dynamic indicator of changes in body water content. A relatively pronounced negative fluid balance was required for a significant corresponding directional change in BIVA hydration to occur. Moreover, the correlation between the overall change in BIVA hydration and the corresponding directional change in fluid balance was weak and of borderline significance. These conflicting findings support the need for further investigations of the utility of BIVA in critically ill patients.

Key messages

  • BIVA-measured hydration is an additional tool for fluid status assessment in critically ill patients and is feasible to use in such patients.

  • Directional changes in BIVA were consistent with directional changes in fluid balance.

  • BIVA measurements of hydration status correlated with cumulative fluid balance during the time in the ICU.

  • The findings support that BIVA hydration is a valid measure of fluid status and justifies further studies to assess the utility of BIVA in critically ill patients.

Abbreviations

APACHE III:

Acute Physiology and Chronic Health Evaluation

BIVA:

Bioelectrical impedance vector analysis

COPD:

Chronic obstructive pulmonary disease

CRRT:

Continuous renal replacement therapy

CVP:

Central venous pressure

FiO2 :

Fraction of inspired oxygen

H:

Height

ICU:

Intensive care unit

IQR:

Interquartile range

MAP:

Mean arterial pressure

NT-pro-BNP:

N-terminal pro-B-type natriuretic peptide

PaO2 :

Partial pressure of arterial oxygen

R:

Resistance

RM-ANOVA:

Repeated-measures analysis of variance

TBW:

Total body water

Xc:

Reactance

References

  1. Marik PE, Baram M, Vahid B. Does central venous pressure predict volume responsiveness? A systematic review of the literature and the tale of seven mares. Chest. 2008;134:172–8.

    Article  Google Scholar 

  2. Kumar A, Anel R, Bunnell E, Habet K, Zanotti S, Marshall S, et al. Pulmonary artery occlusion pressure and central venous pressure fail to predict ventricular filling volume, cardiac performance, or the response to volume infusion in normal subjects. Crit Care Med. 2004;32:691–9.

    Article  Google Scholar 

  3. Michard F, Teboul JL. Predicting fluid responsiveness in ICU patients: a critical analysis of the evidence. Chest. 2002;121:2000–8.

    Article  Google Scholar 

  4. Perren A, Markmann M, Merlani G, Marone C, Merlani P. Fluid balance in critically ill patients: should we really rely on it? Minerva Anesthesiol. 2011;77:802–11.

    CAS  Google Scholar 

  5. Schneider A, Baldwin I, Freitag E, Glassford N, Bellomo R. Estimation of fluid status changes in critically ill patients: fluid balance or electronic bed weight? J Crit Care. 2012;27:745.e7–e12.

    Article  Google Scholar 

  6. Schneider AG, Thorpe C, Delbridge K, Matalanis G, Bellomo R. Electronic bed-weighing vs daily fluid balance changes after cardiac surgery. J Crit Care. 2013;28:1113.e1–e5.

    Article  Google Scholar 

  7. Stevenson LW, Perloff JK. The limited availability of physical signs for estimating hemodynamics in chronic heart failure. JAMA. 1989;261:884–8.

    Article  CAS  Google Scholar 

  8. Di Somma S, Navarin S, Giordano S, Spadini F, Lippi G, Cervellin G, et al. The emerging role of biomarkers and bioimpedance in evaluating hydration status in patients with acute heart failure. Clin Chem Lab Med. 2012;50:2093–105.

    Article  Google Scholar 

  9. Kushner RF, Schoeller GA, Fjeld CR, Danford L. Is the impedance index (ht2/R) significant in predicting total body water? Am J Clin Nutr. 1992;56:835–9.

    Article  CAS  Google Scholar 

  10. Lukaski HC, Hall CB, Siders WA. Assessment of hydration status in women during pregnancy and postpartum with bioelectrical impedance vectors. Nutrition. 2007;23:543–50.

    Article  Google Scholar 

  11. Ronco C, Kaushik M, Valle R, Aspromonte N, Peacock 4th WF. Diagnosis and management of fluid overload in heart failure and cardio-renal syndrome: the “5B” approach. Semin Nephrol. 2012;32:129–41.

    Article  Google Scholar 

  12. Uszko-Lencer N, Bothmer F, Van Pol PE, Schols AM. Measurement of body composition in chronic heart failure: a comparison of methods. Eur J Heart Fail. 2006;8:208–14.

    Article  Google Scholar 

  13. Piccoli A. the Italian Hemodialysis-Bioelectrical Impedance Analysis (HD-BIA) Study Group. Identification of operational clues to dry weight prescription in haemodialysis using bioimpedance vector analysis. Kidney Int. 1998;53:1036–43.

    Article  CAS  Google Scholar 

  14. Jha V, Jairam A, Sharma MC, Sakhuja V, Piccoli A, Parthasarathy S. Body composition analysis with bioelectric impedance in adult Indians with ESRD: comparison with healthy population. Kidney Int. 2006;69:1649–53.

    Article  CAS  Google Scholar 

  15. Pillon L, Piccoli A, Lowrie EG, Lazarus JM, Chertow GM. Vector length as a proxy for the adequacy of ultrafiltration in haemodialysis. Kidney Int. 2004;66:1266–71.

    Article  Google Scholar 

  16. Onofriescu M, Hogas S, Voroneanu L, Apetrii M, Nistor I, Kanbay M, et al. Bioimpedance-guided fluid management in maintenance haemodialysis: a pilot randomised controlled trial. Am J Kidney Dis. 2014;64:111–8.

    Article  Google Scholar 

  17. Piccoli A, Condognotto M, Pasesntin P, Naso A. Combined evaluation of nutrition and hydration in dialysis patients with bioelectrical impedance vector analysis (BIVA). Clin Nutr. 2014;33:673–7.

    Article  Google Scholar 

  18. Peacock 4th WF. Use of bioimpedance vector analysis in critically ill and cardiorenal patients. Contrib Nephrol. 2010;165:226–35.

    Article  Google Scholar 

  19. Piccoli A. Bioelectric impedance measurement for fluid status assessment. Contrib Nephrol. 2010;164:143–52.

    Article  Google Scholar 

  20. Piccoli A, Pittoni G, Facco E, Favaro E, Pillon L. Relationship between central venous pressure and bioimpedance vector analysis in critically ill patients. Crit Care Med. 2000;28:132–7.

    Article  CAS  Google Scholar 

  21. Basso F, Berdin G, Virzi GM, Mason G, Piccinni P, Day S, et al. Fluid management in the intensive care unit: bioelectrical impedance vector analysis as a tool to assess hydration status and optimal fluid balance in critically ill patients. Blood Purif. 2013;36:192–9.

    Article  Google Scholar 

  22. Chen H, Buyun W, Gong D, Liu Z. Fluid overload at start of continuous renal replacement therapy is associated with poorer clinical condition and outcome: a prospective observational study on the combined use of bioimpedance vector analysis and serum N-terminal pro-B-type natriuretic peptide measurement. Crit Care. 2015;19:135.

    Article  Google Scholar 

Download references

Acknowledgements

We thank EFG Diagnostics for providing the Renal EFG BIVA™ Technology device for use in this study. We also thank our ICU nursing staff and physician colleagues and the patients and their families for their support of this study.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Rinaldo Bellomo.

Additional information

Competing interests

The authors declare that they have no competing interests.

Authors’ contributions

SJ, JM and AT obtained BIVA measurements, collected other study data and drafted the manuscript. GE contributed to the design of the study, obtained BIVA measurements, collected other study data and drafted the manuscript. HY and LP collected study data and drafted the first version of the manuscript. RB conceived of the study, participated in its design and coordination and helped to draft the first version of the manuscript. JM performed the statistical analysis. All authors participated in the writing of the final manuscript and read and approved the final manuscript.

Additional file

Additional file 1: Figure S1.

Fluid management. Figure S2. BIVA graphs in CRRT and ECMO patients. Assessment of repeatability. (DOCX 564 kb)

Rights and permissions

Open Access This article is distributed under the terms of the Creative Commons Attribution 4.0 International License (http://creativecommons.org/licenses/by/4.0/), which permits unrestricted use, distribution, and reproduction in any medium, provided you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons license, and indicate if changes were made. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated.

Reprints and permissions

About this article

Check for updates. Verify currency and authenticity via CrossMark

Cite this article

Jones, S.L., Tanaka, A., Eastwood, G.M. et al. Bioelectrical impedance vector analysis in critically ill patients: a prospective, clinician-blinded investigation. Crit Care 19, 290 (2015). https://doi.org/10.1186/s13054-015-1009-3

Download citation

  • Received:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/s13054-015-1009-3

Keywords