In vitro efficiency of Amikacin Inhale, a novel drug-device delivery system

Introduction

Amikacin Inhale (NKTR-061, BAY41-6551) is a drug-device combination in clinical development for adjunctive treatment of intubated and mechanically ventilated patients with Gram-negative pneumonia. The product uses a proprietary vibrating mesh nebulizer system (PDDS Clinical) with amikacin sulfate formulated for inhalation (3.2 ml of 125 mg/ml amikacin solution) for a 10-day twice-daily course of therapy. It is designed for use with two delivery systems: one system for intubated patients (On-vent; Figure 1), and a second Handheld (HH) system for patients who are extubated before completing the course of therapy (Figure 2). We investigated in vitro the amikacin lung dose delivered by PDDS Clinical.

Figure 1
figure 1

On-vent system with brown blinder shell for clinical trials.

Figure 2
figure 2

Handheld system with brown blinder shell for clinical trials.

Methods

An estimated lung dose (ELD) for On-vent setting was measured in vitro after collecting aerosolized amikacin from a filter at the end of an endotracheal tube during ventilation. The ELD for the HH device was calculated from the fine particle fraction (FPF <5 µm) post-mouthpiece, multiplied by the in vitro delivered dose post-mouthpiece. FPF <5 µm reflects lung deposition observed during phase 2 clinical trials [1]. Eighty-one nebulizers with volume median diameter (VMD) of 4.4 ± 0.5 µm and output rates of 0.23 ± 0.10 ml/minute were tested for each system. Delivered dose data were fit to the independent variables (that is, VMD and output rates) using a least-squares fit with 95% confidence limits.

Results

Total percentage recoveries for On-vent and HH test runs were between 85% and 115% of the nominal dose. The mean ELDs were 50 ± 9% (On-vent) and 49 ± 11% (HH) of the nominal dose. Nebulizers with longer dosing times and lower VMDs had higher ELD values for both delivery systems.

Conclusion

The results support the use of the PDDS Clinical with either system to administer aerosolized amikacin with high efficiency and no dose adjustment is required when switching from the On-vent to the HH system for extubated patients.

References

  1. Fink J, et al.: High in vivo amikacin lung deposition after NKTR-061 dosing correlates with in vitro aerosol characterization. Am J Respir Crit Care Med 2008, 177: A530.

    Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Rights and permissions

This article is licensed under a Creative Commons Attribution 4.0 International License, which permits use, sharing, adaptation, distribution and reproduction in any medium or format, as long as you give appropriate credit to the original author(s) and the source, provide a link to the Creative Commons licence, and indicate if changes were made. The images or other third party material in this article are included in the article's Creative Commons licence, unless indicated otherwise in a credit line to the material. If material is not included in the article's Creative Commons licence and your intended use is not permitted by statutory regulation or exceeds the permitted use, you will need to obtain permission directly from the copyright holder. To view a copy of this licence, visit http://creativecommons.org/licenses/by/4.0/. The Creative Commons Public Domain Dedication waiver (http://creativecommons.org/publicdomain/zero/1.0/) applies to the data made available in this article, unless otherwise stated in a credit line to the data.

Reprints and permissions

About this article

Cite this article

Kadrichu, N., Boc, S., Corkery, K. et al. In vitro efficiency of Amikacin Inhale, a novel drug-device delivery system. Crit Care 17 (Suppl 2), P81 (2013). https://doi.org/10.1186/cc12019

Download citation

  • Published:

  • DOI: https://doi.org/10.1186/cc12019

Keywords