Skip to main content

Withdrawing intra-aortic balloon pump support paradoxically improves microvascular flow

Abstract

Introduction

The Intra-Aortic Balloon Pump (IABP) is frequently used to mechanically support the heart. There is evidence that IABP improves microvascular flow during cardiogenic shock but its influence on the human microcirculation in patients deemed ready for discontinuing IABP support has not yet been studied. Therefore we used sidestream dark field imaging (SDF) to test our hypothesis that human microcirculation remains unaltered with or without IABP support in patients clinically ready for discontinuation of mechanical support.

Methods

We studied 15 ICU patients on IABP therapy. Measurements were performed after the clinical decision was made to remove the balloon catheter. We recorded global hemodynamic parameters and performed venous oximetry during maximal IABP support (1:1) and 10 minutes after temporarily stopping the IABP therapy. At both time points, we also recorded video clips of the sublingual microcirculation. From these we determined indices of microvascular perfusion including perfused vessel density (PVD) and microvascular flow index (MFI).

Results

Ceasing IABP support lowered mean arterial pressure (74 ± 8 to 71 ± 10 mmHg; P = 0.048) and increased diastolic pressure (43 ± 10 to 53 ± 9 mmHg; P = 0.0002). However, at the level of the microcirculation we found an increase of PVD of small vessels <20 μm (5.47 ± 1.76 to 6.63 ± 1.90; P = 0.0039). PVD for vessels >20 μm and MFI for both small and large vessels were unaltered. During the procedure global oxygenation parameters (ScvO2/SvO2) remained unchanged.

Conclusions

In patients deemed ready for discontinuing IABP support according to current practice, SDF imaging showed an increase of microcirculatory flow of small vessels after ceasing IABP therapy. This observation may indicate that IABP impairs microvascular perfusion in recovered patients, although this warrants confirmation.

Introduction

In cardiogenic shock, intra-aortic balloon counterpulsation is frequently used to mechanically support the failing heart [1, 2]. Intra-Aortic Balloon Pump- (IABP-) support improves coronary blood flow by augmenting systemic and coronary diastolic blood pressure and increases cardiac index by reducing left ventricular work [1, 3]. As a bridge to recovery, its goal is to facilitate the heart and continuously provide adequate systemic perfusion.

However, the microcirculation is ultimately responsible for delivering oxygen and substrates to tissue [4]. The recent emergence of Orthogonal Polarization Spectral imaging and its successor sidestream dark field imaging (SDF) has enabled imaging of the human microcirculation in real time [57]. These techniques have been used to characterize the microcirculation in various clinical situations including cardiogenic shock.

For example, De Backer et al.[8] demonstrated that microvascular blood flow worsens in severe cardiac failure and cardiogenic shock and is associated with in-hospital mortality. Importantly, there is now a large body of evidence that microvascular flow may be relatively independent from global hemodynamics [4]. For example, arterial and venous blood pressure, cardiac output as well as central or mixed venous oxygen saturation may not necessarily reflect microvascular perfusion [914]. However, in current clinical practice, these very global hemodynamic parameters frequently guide the decision when to start and withdraw IABP-therapy.

Three recent trials examined the microcirculatory effects of counterpulsation during cardiogenic shock and high risk percutaneous coronary intervention (PCI) [1517]. Two of these reported IABP-induced improvement in microvascular flow whereas the other did not. Therefore our understanding of microvascular perfusion during IABP-support remains based on limited and conflicting data. This paucity of data is even more apparent in deciding when to best withdraw IABP-support. No previous study has addressed this issue. To examine the influence of IABP-support on microcirculation of recovered patients, we studied patients deemed ready for discontinuing IABP-support as judged by their treating physicians. We used SDF-imaging to test our hypothesis that microvascular flow is unaltered with or without IABP-support in this clinical setting.

Materials and methods

The local institutional review board approved the study protocol. Since the study was observational and given the non-invasive nature of SDF-imaging, the need for a written informed consent was waived in accordance with the national Law on Experiments with Humans. The study was performed at the intensive care unit (ICU) of a large teaching hospital between April 2007 and October 2008.

We included adult patients that had an IABP in place. Patients were only included if and when the responsible ICU physician had made the decision that the subjects were clinically ready for discontinuation of IABP-support. We excluded patients that showed signs of sepsis (suspected or proven systemic infection with ≥2 severe inflammatory response syndrome (SIRS) criteria: tachycardia >90 beats/minute and/or tachypnoea >20 breaths/minute or arterial pCO2 < 32 mmHg/4,2kPa and/or body temperature >38°C or <36°C and/or WBC count >12,000 cells/mm3 or <4,000 cells/mm3 or >10% immature cells). Disruption or laceration of the oral floor mucosa was an exclusion criterion because this would interfere with microcirculatory imaging. As per clinical routine, all patients underwent continuous invasive monitoring of arterial and central venous blood pressure and some patients had a surgically placed pulmonary artery catheter.

A Datascope® CS300 intra aortic balloon pump system (Datascope Corporation, Mahwah, NJ, USA) was used in all studied patients. The IABP system was set automatically, using the electrocardiogram for optimal timing so that inflation and deflation occurred at the dicrotic notch and immediately before systolic upstroke, respectively. Optimal balloon size was chosen depending on patient height before insertion. Routine chest X-rays were examined to define correct intra-aortic placement of the IABP balloon, 2 to 3 cm distal to the origin of the left subclavian artery. All hemodynamic parameters were recorded continuously by our patient data management system.

The decision to discontinue IABP-support was a clinical one and left completely at the discretion of the ICU-team. In most cases, this included a weaning trial in which the IABP-assist ratio was lowered step by step over several hours. Possible changes in routinely measured macrocirculatory and laboratory parameters were observed during this process.

Microcirculatory measurements were performed using SDF imaging, which has been described in detail elsewhere [7]. In brief, it consists of a handheld video microscope that emits stroboscopic green light (530 nm) from an outer ring at the tip of the probe. This light is absorbed by haemoglobin. A negative image of moving red blood cells is sent back through the isolated optical core of the probe toward a charge-coupled device (CCD) camera. SDF imaging has been shown to provide a higher imaging quality with more detail and less motion blur than its predecessor Orthogonal Polarization Spectral (Ops) imaging [7]. A typical example of a SDF-image is shown in Figure 1.

Figure 1
figure 1

A screenshot from a typical example of the human sublingual microcirculation using SDF imaging.

Within two hours after the decision to discontinue IABP-support had been made, we performed SDF imaging at two points in time. First, the IABP device was set to a 1:1 assist ratio, if this was not already the selected mode. After 10 minutes, to allow for a new steady state to occur, the first series of microvascular recordings was made. Next, the IABP device was temporarily stopped. After another 10 minutes, we recorded a second series of SDF video clips. At both measuring points, venous and arterial blood gas analyses were collected. In patients with a pulmonary artery catheter mixed venous saturation (SvO2) was determined, otherwise central venous saturation (ScvO2) was measured. After the procedure the IABP device was switched back on at the pre-measurement settings. During the procedure dosages of continuous intravenous drugs were recorded and no dosing adjustments were made.

In a recent round table conference, international experts reached consensus on how to best evaluate the microcirculation using OPS and SDF imaging [18]. We implemented all recommendations given in this conference. Video clips were immediately saved as digital AVI-DV files to a hard drive of a personal computer using an analogue-to-digital converter (Canopus, Kobe, Japan) and the freeware program WinDV [19]. We used 5× optical magnification, producing images representing approximately 940 × 750 mm2 of tissue surface. Per measuring point, clips at three sublingual sites yielding at least 20 seconds of stable video per site were recorded. Special care was taken to avoid pressure artefacts, adhering to the standard operating procedure previously described by Trzeciak et al.[20] and recommended in the round table conference [18]. In brief, secretions were removed with gauze, and, after obtaining good imaging focus, the probe was pulled back gently until contact was lost and then advanced again slowly to the point at which contact was regained. The authors paid special attention to the larger vessels at the time of recording because alterations in their flow with probe manipulation may indicate pressure artefacts.

One video file was recorded for each location at each measuring point. These were stored under a random number. At a later time, these were analyzed by one of the authors (PWGE) using the AVA 3.0 software program (Microvision Medical, Amsterdam, The Netherlands). According to the recommendations, microvascular flow index (MFI), perfused vessel density (PVD), proportion of perfused vessels (PPVs), and indices of heterogeneity were determined for every patient at both time points. All have been validated previously [2022]. As published recently, each score was determined for both large and small microvessels, with a cut-off diameter of 20 μm [23]. In addition, the authors defined large-type vessels that split into other vessels as arterioles. Other large vessels were defined as venules.

For PPV and PVD, vessel density was calculated as the number of vessels crossing three horizontal and three vertical equidistant lines spanning the screen divided by the total length of the lines. Perfusion at each crossing was then scored semi-quantitatively by the eye as follows: 0 = no flow (no flow present for the entire duration of the clip), 1 = intermittent flow (flow present <50% of the duration of the clip), 2 = sluggish flow (flow present >50% but <100% of the duration of the clip or very slow flow for the entire duration of the clip), and 3 = continuous flow (flow present for the entire duration of the clip). PVD was then calculated as the number of crossings with flow scores greater than 1. PPV was calculated as the proportion of crossings with flow scores greater than 1 divided by the total number of crossings. For each measuring point and each patient, the scores for PPV and PVD were averaged. PVD is expressed in n/mm, whereas PPV is expressed as a percentage. Intra-observer variability ranges between 2.5% and 4.7% for PPV and between 0.9% and 4.5% for PPV. The inter-observer variability is slightly higher: between 3.0% and 6.2% and between 4.1% and 10%, respectively [21].

MFI was based on the determination of the predominant type of flow in four quadrants adhering to the same scoring system. MFI is the sum of these flow scores divided by the number of quadrants in which the vessel type is visible. The intra-observer agreement of MFI is about 85% (kappa score = 0.78) and inter-observer agreement about 90% (kappa score = 0.85) [22]. For each measuring point and each patient, the scores for MFI were averaged.

Heterogeneity was assessed in two different ways. For PVD, the coefficient of variation was determined. For MFI, the authors assessed heterogeneity in each patient by subtracting the lowest from the highest quadrant MFI and dividing the result by the mean MFI [20, 21].

We used Wilcoxon matched pairs tests for MFI and paired t tests for other data. We used Spearman tests to detect correlation between global and microvascular parameters. Results are reported as median and interquartile ranges (IQR) for MFI and as the mean ± standard deviation (SD) for other parameters, unless indicated otherwise. The study was powered to detect a minimum of 15% difference in small-vessel PVD after switching off the IABP with α = 0.05 and 1-β = 0.80. Based on previous studies by others and us [8, 10, 21] this showed the need for inclusion of 14 patients.

Results

Patients

We included 15 patients. Baseline characteristics of the study population, including risk factors for cardiovascular disease and continuous intravenous dosage of vasoactive drugs, are shown in Table 1. All participants were admitted at the intensive care unit of St. Antonius Hospital, Nieuwegein, The Netherlands.

Table 1 Patient characteristics and drugs

Mean APACHE (acute physiology and chronic health evaluation scoring system) II score 24 hours after admission at the ICU was 14.5 ± 5.3. Indications for IABP placement were urgent coronary artery bypass grafting due to unstable angina pectoris with severe coronary disease (n = 6, 40.0%), cardiogenic shock due to myocardial infarction without heart surgery (n = 2, 13.3%), cardiogenic shock after heart surgery (n = 2, 13.3%), due to acute prosthetic aortic valve displacement after surgery (n = 1, 6.7%) or after cardiac arrest with or without percutaneous coronary intervention (n = 4, 26.6%). Mean duration of IABP-therapy at the time of measurement was three days (range one to five days). During the experiment, six patients had a pulmonary artery catheter in place and 11 patients were mechanically ventilated.

Systemic hemodynamic data

Table 2 depicts the differences in global hemodynamic parameters between the two points of interest. The institution of IABP-support significantly increased mean arterial pressure (MAP). However, recorded diastolic blood pressure was significantly lower. It is important to point out that this represents the lowest pressure recorded during the cardiac cycle, which is the purpose of balloon counterpulsation. After switching off IABP-support no differences in venous oxygen saturation occurred (ScvO2/SvO2; 70.9% ± 7.2 vs. 71.4% ± 7.5; P = 0.897).

Table 2 Global hemodynamic data

Microcirculation

The authors successfully obtained high-quality images in each patient. In total 90 video clips were recorded. Results are shown in Table 3. PVD of small vessels (< 20 μm) was significantly lower during IABP-support; 5.47 ± 1.76 vs. 6.63 ± 1.90; P = 0.0039 (Figure 2). Other microcirculatory parameters were not significantly altered.

Table 3 Microcirculation
Figure 2
figure 2

The effect of IABP on perfused vessel density (PVD) for vessels <20 μm. The perfused vessel density increased significantly, 5.47 ± 1.76 to 6.63 ± 1.90 (P = 0.0039), 10 minutes after IABP switch-off.

The relationship between changes in mean arterial pressure, diastolic arterial pressure and S(c)vO2 versus changes in PVD in individual patients comparing maximal support vs. no support was determined. Spearman tests did not show a statistically significant correlation for any of these global hemodynamic parameters (r = -0.1; P = 0.71, r = 0.1; P = 0.72, r = 0.1; P = 0.69 respectively). In addition, no correlation was found between small vessel PVD and APACHE II score (r = 0.04; P = 0.88).

Discussion

To the best of our knowledge, this is the first study to report on the microvascular effects of IABP switch-off in recovered patients deemed ready for IABP removal. Our most prominent finding is that cessation of IABP-support resulted in a significant and paradoxical increase in small-vessel PVD, independent from initial disease severity. This is in contrast with the loss of IABP-induced raise of mean arterial pressure in these patients [13] and, perhaps more importantly, in contrast with SvO2 or ScvO2 values, which remained unchanged after IABP switch-off. Therefore, this study adds to the large body of evidence that global hemodynamic parameters and venous oximetry do not necessarily reflect microvascular perfusion [4, 9, 10, 21, 22]. It is the first time that such a discrepancy is demonstrated after withdrawal of IABP-support.

Three recent studies examined the effects of IABP on human microcirculation [1517]. Jung et al. included 13 patients with cardiogenic shock after acute myocardial infarction. The authors recorded SDF video images before and shortly after the IABP-support was temporarily stopped. MFI of small and medium vessels (10 to 50 μm) was significantly higher in patients with IABP-support [15]. In contrast, den Uil et al. studied a heterogeneous group of 13 patients suffering from cardiogenic shock of variable severity and found no differences in perfused capillary density (PCD) and red blood cell velocity [16]. This was despite the fact that mean arterial pressure and cardiac index were significantly lower after the IABP-assist ratio was switched from 1:1 to 1:8. Finally, more recently, Jung et al. studied six patients immediately following high-risk PCI. MFI of both small and large vessels decreased significantly immediately after a short period of IABP-support discontinuation and returned to baseline after restarting therapy. Again, no correlation with global hemodynamic parameters was found [17].

Both groups chose not to fully incorporate recent recommendations regarding image acquisition or reporting standard microvascular flow parameters [18]. This hampers direct comparison with our study. More importantly, our population of recovered patients is markedly different. However, this does not dismiss our sharply contrasting finding of improved PVD after IABP switch-off. It is plausible that IABP has different microvascular effects at different stages of disease and recovery.

Although speculative, one possible explanation for the observed difference in microvascular perfusion could be a mechanical effect of IABP. The pressure decrease induced by balloon deflation during diastole could create a collapse of certain parts of capillary microcirculation. This effect is consistent with the concept of vascular waterfalls: the flow in front of a waterfall is not affected by its height. This analogy explains why sometimes global and regional blood flow ceases at positive arterial pressures well above venous pressure [24, 25]. Interestingly, an IABP-induced transient blood flow reversal in basal cerebral arteries has been reported previously [26, 27], which is consistent with this theory. However, correlation between the difference in diastolic pressure and the difference in PVD between the study time points did not reach statistical significance. Conversely, this may not invalidate our hypothesis, as microvascular de-recruitment is probably an all or nothing stochastic process.

There are several important limitations to our study. First, we did not routinely measure cardiac output. However, this is highly representative of current practice. In addition, it may be reasonable to assume that cardiac output remained constant as no changes in venous oximetry and heart rate were observed and no drug dosing adjustments were made during the study period. Interestingly, if IABP would have increased cardiac output in our patients, as it has been reported to do so in other clinical situations, this would perhaps even strengthen our findings. This would mean that despite increased cardiac output during IABP-support, microvascular flow would be relatively impaired as compared to no IABP-support.

Second, we chose the sublingual site for its proximity to the brain, ease of access and its philogenetic relationship to the gut. There is conflicting data on whether sublingual microvascular perfusion represents other vascular beds [28, 29]. Interestingly, a recent animal study shows that in the setting of resuscitation, cerebral microcirculation is relatively protected as compared to sublingual microvascular perfusion [30]. If applicable to the setting studied by us, this would imply that cerebral microvascular perfusion improves to a greater extent after IABP switch-off. Third, the fact that our findings could possibly be explained by temporal changes in microcirculation instead of IABP cannot be ruled out because we did not perform a final measurement after reinstituting IABP-support. Finally, but perhaps most important, a heterogeneous group was studied in which a minority initially had cardiogenic shock. However, no correlation of initial disease severity (APACHE II score) and microcirculatory perfusion at the time of measurement could be found. Further, there were no differences in microcirculatory perfusion between patients with initial shock and those without shock. The absence of correlation between microvascular flow and APACHE II and shock state may not be surprising as patients were only studied during a phase in which they were deemed ready for discontinuation of IABP-support.

Given these limitations, our findings and their clinical consequences should be interpreted with caution. In addition, while we have shown that IABP may impair tissue perfusion in hemodynamically recovered patients, we did not observe signs of cell ischemia (for example, lactate acidosis) nor signs of progressive organ dysfunction during IABP-therapy on the day measurements were performed. In addition PVD values during IABP-support are similar to those found in healthy controls by De Backer [8] and to those found after routine cardiac surgery by us [10].

However, given the current controversy on the evidence of IABP-support and therefore its indication [31], coupled with complication risks [32], it may be prudent not to ignore our results. Perhaps the best strategy is to optimize the duration of support and our findings could possibly be the consequence of the fact that our patients may have been too long IABP-treated. This could explain our findings. The role of microvascular imaging to achieve this goal merits further study.

Conclusions

The results of this study show that cessation of IABP-support resulted in a paradoxical increase of microvascular flow in small vessels at a time clinicians deemed IABP was no longer necessary. These changes are independent from global hemodynamic parameters and oxygen derived variables.

Clinicians that routinely rely on these parameters for their decision-making should again be reminded that these do not necessarily reflect microvascular perfusion in a wide range of clinical settings including that of cessation of IABP-support.

Key messages

  • Discontinuing IABP-therapy increases sublingual microvascular perfusion of small vessels in patients deemed clinically ready for ceasing IABP-support.

  • Global hemodynamic parameters and venous oximetry do not correlate with and may therefore be unreliable for predicting adequacy of microvascular perfusion in this clinical setting.

Abbreviations

APACHE:

acute physiology and chronic health evaluation scoring system

CCD:

charge-coupled device

IABP:

intra aortic balloon pump

MAP:

mean arterial pressure

MFI:

microvascular flow index

OPS:

orthogonal polarization spectral imaging

PCD:

perfused capillary density

PCI:

percutaneous coronary intervention

PPV:

proportion of perfused vessels

PVD:

perfused vessel density

ScvO2:

central venous oxygen saturation

SD:

standard deviation

SDF:

sidestream dark field imaging

SvO2:

mixed venous oxygen saturation.

References

  1. Trost JC, Hillis LD: Intra-aortic balloon counterpulsation. Am J Cardiol 2006, 97: 1391-1398. 10.1016/j.amjcard.2005.11.070

    Article  PubMed  Google Scholar 

  2. Santa-Cruz RA, Cohen MG, Ohman EM: Aortic counterpulsation: a review of the hemodynamic effects and indications for use. Catheter Cardiovasc Interv 2006, 67: 68-77. 10.1002/ccd.20552

    Article  PubMed  Google Scholar 

  3. Nanas JN, Moulopoulos SD: Counterpulsation: historical background, technical improvements, hemodynamic and metabolic effects. Cardiology 1994, 84: 156-167. 10.1159/000176394

    Article  CAS  PubMed  Google Scholar 

  4. Elbers PW, Ince C: Mechanisms of critical illness--classifying microcirculatory flow abnormalities in distributive shock. Crit Care 2006, 10: 221. 10.1186/cc4969

    Article  PubMed Central  PubMed  Google Scholar 

  5. Mathura KR, Vollebregt KC, Boer K, De Graaff JC, Ubbink DT, Ince C: Comparison of OPS imaging and conventional capillary microscopy to study the human microcirculation. J Appl Physiol 2001, 91: 74-78.

    CAS  PubMed  Google Scholar 

  6. Groner W, Winkelman JW, Harris AG, Ince C, Bouma GJ, Messmer K, Nadeau RG: Orthogonal polarization spectral imaging: a new method for study of the microcirculation. Nat Med 1999, 5: 1209-1212. 10.1038/13529

    Article  CAS  PubMed  Google Scholar 

  7. Goedhart PT, Khalilzada M, Bezemer R, Merza J, Ince C: Sidestream Dark Field (SDF) imaging: a novel stroboscopic LED ring-based imaging modality for clinical assessment of the microcirculation. Opt Express 2007, 15: 15101-15114. 10.1364/OE.15.015101

    Article  CAS  PubMed  Google Scholar 

  8. De Backer D, Creteur J, Dubois MJ, Sakr Y, Vincent JL: Microvascular alterations in patients with acute severe heart failure and cardiogenic shock. Am Heart J 2004, 147: 91-99. 10.1016/j.ahj.2003.07.006

    Article  PubMed  Google Scholar 

  9. den Uil CA, Klijn E, Lagrand WK, Brugts JJ, Ince C, Spronk PE, Simoons ML: The microcirculation in health and critical disease. Prog Cardiovasc Dis 2008, 51: 161-170. 10.1016/j.pcad.2008.07.002

    Article  PubMed  Google Scholar 

  10. Elbers PW, Ozdemir A, van Iterson M, van Dongen EP, Ince C: Microcirculatory imaging in cardiac anesthesia: ketanserin reduces blood pressure but not perfused capillary density. J Cardiothorac Vasc Anesth 2009, 23: 95-101. 10.1053/j.jvca.2008.09.013

    Article  PubMed  Google Scholar 

  11. De Backer D, Creteur J, Dubois MJ, Sakr Y, Koch M, Verdant C, Vincent JL: The effects of dobutamine on microcirculatory alterations in patients with septic shock are independent of its systemic effects. Crit Care Med 2006, 34: 403-408. 10.1097/01.CCM.0000198107.61493.5A

    Article  CAS  PubMed  Google Scholar 

  12. Ince C, Sinaasappel M: Microcirculatory oxygenation and shunting in sepsis and shock. Crit Care Med 1999, 27: 1369-1377. 10.1097/00003246-199907000-00031

    Article  CAS  PubMed  Google Scholar 

  13. Klijn E, den Uil CA, Bakker J, Ince C: The heterogeneity of the microcirculation in critical illness. Clin Chest Med 2008, 29: 643-654. 10.1016/j.ccm.2008.06.008

    Article  PubMed  Google Scholar 

  14. Dubin A, Pozo MO, Casabella CA, Palizas F Jr, Murias G, Moseinco MC, Kanoore Edul VS, Palizas F, Estenssoro E, Ince C: Increasing arterial blood pressure with norepinephrine does not improve microcirculatory blood flow: a prospective study. Crit Care 2009, 13: R92. 10.1186/cc7922

    Article  PubMed Central  PubMed  Google Scholar 

  15. Jung C, Rödiger C, Fritzenwanger M, Schumm J, Lauten A, Figulla HR, Ferrari M: Acute microflow changes after stop and restart of intra-aortic balloon pump in cardiogenic shock. Clin Res Cardiol 2009, 98: 469-475. 10.1007/s00392-009-0018-0

    Article  CAS  PubMed  Google Scholar 

  16. den Uil CA, Lagrand WK, van der Ent M, Jewbali LS, Brugts JJ, Spronk PE, Simoons ML: The effects of intra-aortic balloon pump support on macrocirculation and tissue microcirculation in patients with cardiogenic shock. Cardiology 2009, 114: 42-46. 10.1159/000212060

    Article  CAS  PubMed  Google Scholar 

  17. Jung C, Lauten A, Rödiger C, Krizanic F, Figulla HR, Ferrari M: Effect of intra-aortic balloon pump support on microcirculation during high-risk percutaneous intervention. Perfusion 2009, 24: 417-421. 10.1177/0267659109358208

    Article  PubMed  Google Scholar 

  18. De Backer D, Hollenberg S, Boerma C, Goedhart P, Buchele G, Ospina-Tascon G, Dobbe I, Ince C: How to evaluate the microcirculation: report of a round table conference. Crit Care 2007, 11: R101. 10.1186/cc6118

    Article  PubMed Central  PubMed  Google Scholar 

  19. WinDV, a small Windows utility for DV (FireWire digital video) input/output[http://windv.mourek.cz/]

  20. Trzeciak S, Dellinger RP, Parrillo JE, Guglielmi M, Bajaj J, Abate NL, Arnold RC, Colilla S, Zanotti S, Hollenberg SM: Microcirculatory Alterations in Resuscitation and Shock Investigators. Early microcirculatory perfusion derangements in patients with severe sepsis and septic shock: relationship to hemodynamics, oxygen transport, and survival. Ann Emerg Med 2007, 49: 88-98. 10.1016/j.annemergmed.2006.08.021

    Article  PubMed  Google Scholar 

  21. De Backer D, Creteur J, Preiser JC, Dubois MJ, Vincent JL: Microvascular blood flow is altered in patients with sepsis. Am J Respir Crit Care Med 2002, 166: 98-104. 10.1164/rccm.200109-016OC

    Article  PubMed  Google Scholar 

  22. Boerma EC, Mathura KR, van der Voort PH, Spronk PE, Ince C: Quantifying bedside-derived imaging of microcirculatory abnormalities in septic patients: a prospective validation study. Crit Care 2005, 9: R601-606. 10.1186/cc3809

    Article  PubMed Central  PubMed  Google Scholar 

  23. Elbers PW: Fast-track microcirculation analysis. Crit Care 2007, 11: 426. 10.1186/cc6182

    Article  PubMed Central  PubMed  Google Scholar 

  24. Hoffman JI, Spaan JA: Pressure-flow relations in coronary circulation. Physiol Rev 1990, 70: 331-390.

    CAS  PubMed  Google Scholar 

  25. Magder S: Starling resistor versus compliance. Which explains the zero-flow pressure of a dynamic arterial pressure-flow relation? Circ Res 1990, 67: 209-220.

    Article  CAS  PubMed  Google Scholar 

  26. Schachtrupp A, Wrigge H, Busch T, Buhre W, Weyland A: Influence of intra-aortic balloon pumping on cerebral blood flow pattern in patients after cardiac surgery. Eur J Anaesthesiol 2005, 22: 165-170.

    Article  CAS  PubMed  Google Scholar 

  27. Brass LM: Reversed intracranial blood flow in patients with an intra-aortic balloon pump. Stroke 1990, 21: 484-487.

    Article  CAS  PubMed  Google Scholar 

  28. Boerma EC, van der Voort PH, Spronk PE, Ince C: Relationship between sublingual and intestinal microcirculatory perfusion in patients with abdominal sepsis. Crit Care Med 2007, 35: 1055-1060. 10.1097/01.CCM.0000259527.89927.F9

    Article  PubMed  Google Scholar 

  29. Verdant CL, De Backer D, Bruhn A, Clausi CM, Su F, Wang Z, Rodriguez H, Pries AR, Vincent JL: Evaluation of sublingual and gut mucosal microcirculation in sepsis: a quantitative analysis. Crit Care Med 2009, 37: 2875-2881. 10.1097/CCM.0b013e3181b029c1

    Article  PubMed  Google Scholar 

  30. Wan Z, Sun S, Ristagno G, Weil MH, Tang W: The cerebral microcirculation is protected during experimental hemorrhagic shock. Crit Care Med 2010, 38: 928-932. 10.1097/CCM.0b013e3181cd100c

    Article  PubMed  Google Scholar 

  31. Prondzinsky R, Lemm H, Swyter M, Wegener N, Unverzagt S, Carter JM, Russ M, Schlitt A, Buerke U, Christoph A, Schmidt H, Winkler M, Thiery J, Werdan K, Buerke M: Intra-aortic balloon counterpulsation in patients with acute myocardial infarction complicated by cardiogenic shock: the prospective, randomized IABP SHOCK Trial for attenuation of multiorgan dysfunction syndrome. Crit Care Med 2010, 38: 152-160. 10.1097/CCM.0b013e3181b78671

    Article  CAS  PubMed  Google Scholar 

  32. Sjauw KD, Engstrom AE, Vis MM, van der Schaaf RJ, Baan J Jr, Koch KT, de Winter RJ, Piek JJ, Tijssen JG, Henriques JP: A systematic review and meta-analysis of intra-aortic balloon pump therapy in ST-elevation myocardial infarction: should we change the guidelines? Eur Heart J 2009, 30: 459-468. 10.1093/eurheartj/ehn602

    Article  PubMed  Google Scholar 

Download references

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Luuk DH Munsterman.

Additional information

Competing interests

CI is chief scientific officer of Microvision Medical. Microvision Medical is a university-based company dedicated to the development of optical spectroscopic tools for study of the microcirculation and tissue oxygenation, in which context CI holds patents and shares. All other authors declare they have no competing interests.

Authors' contributions

LM participated in the design of the study, performed data acquisition and drafted the manuscript. PGWE conceived the study, performed data processing and drafting of the manuscript. AO participated in data acquisition. ED, MI and CI have participated in the design of the study and critically revising of the manuscript.

Luuk DH Munsterman, Paul WG Elbers contributed equally to this work.

Authors’ original submitted files for images

Rights and permissions

Reprints and permissions

About this article

Cite this article

Munsterman, L.D., Elbers, P.W., Ozdemir, A. et al. Withdrawing intra-aortic balloon pump support paradoxically improves microvascular flow. Crit Care 14, R161 (2010). https://doi.org/10.1186/cc9242

Download citation

  • Received:

  • Revised:

  • Accepted:

  • Published:

  • DOI: https://doi.org/10.1186/cc9242

Keywords