Generation of a single pulmonary pressure-volume curve does not durably affect oxygenation in patients with acute respiratory distress syndrome

Abstract

Introduction

It is possible that taking a static pressure-volume (PV) measurement could durably affect oxygenation and thus interfere with early evaluation of a therapeutic intervention delivered just after that measurement. The aim of the present study was to investigate the effects over time of a single static PV measurement on gas exchange and haemodynamics; the PV measurements were taken using a super syringe and by using the constant flow method in patients with acute respiratory distress syndrome.

Method

We conducted a prospective, randomized and controlled interventional study in an intensive care unit. The study was conducted in 17 patients with early acute respiratory distress syndrome ventilated with a tidal volume of 6.9 ± 1.0 ml/kg, a plateau pressure of 27 ± 7 cmH2O and a positive end-expiratory pressure [PEEP] of 10 cmH2O. They were all evaluated for 1 hour after each of the following two measurements was taken and during a control period (in a randomized order): generation of a PV curve using a 2 l super syringe (PVSS; insufflated volume = 1824 ± 381 ml, plateau pressure = 46 ± 9 cmH2O); and generation of a PV curve using the constant flow method on the ventilator (PVCF; insufflated volume = 1120 ± 115 ml in zero end-expiratory pressure after 20 s expiratory pause, plateau pressure = 46 ± 11 cmH2O). The maximal airway pressure allowed during PV measurement was 60 cmH2O. PEEP was set to 10 cmH2O immediately after PV measurement. Partial arterial oxygen tension (Pao2), partial carbon dioxide tension (Paco2) and mean arterial pressure were recorded each minute.

Results

PV measurement did not significantly affect Pao2, Paco2, mean arterial pressure and lung mechanics. Two patients exhibited a sustained increase in Pao2 by more than 20% after PVCF (>60 minutes). Two patients exhibited a decrease in Pao2 by more than 20% after PVSS, which was sustained in one. These latter patients had an upper inflection point identified on the PV curve. After PVSS, Paco2 increased by more than 10 mmHg in two patients and returned to baseline values after 15 minutes. One patient exhibited a decrease in mean arterial pressure by more than 10 mmHg for less than 5 minutes after PVSS and one patient after PVCF.

Conclusion

Evaluation of the effects of a strategy aimed at improving oxygenation can be reliably recorded early after a single PV measurement that is not followed by a change in PEEP level. PV measurement using the constant flow method improves oxygenation in a limited number of patients.

Introduction

The pressure-volume (PV) curve characteristics of the respiratory system are commonly evaluated in clinical and experimental studies of acute respiratory distress syndrome (ARDS). The PV measurement involves insufflating the lungs at low flow with a volume of up to 2 l using a super syringe [1] or about 1200 ml by ventilator [2], which is done in order to construct a static PV curve. The procedures required to construct PV curves may improve oxygenation because they result in alveolar recruitment. On the other hand, there are aspects of the two procedures that could result in impaired oxygenation; specifically, it is necessary to disconnect the patient from the ventilator before and after PV curve measurements with the super syringe technique, and with the constant flow method positive end-expiratory pressure (PEEP) must be removed before the PV curve measurements can be taken [3, 4]. However, the potential sustained effects of PV measurement on gas exchange and haemodynamic parameters have not been investigated in patients presenting with acute respiratory distress syndrome (ARDS). This is of concern when ventilator settings (such as adjusting PEEP level) or any other intervention (for example prone positioning) are studied just after PV measurement and evaluated by blood gas analysis during the following 20–60 minutes. In these situations it is important to know how long one should to wait after PV measurement to obtain stable oxygenation parameters. The present study, conducted in ARDS patients, compared the effects over time of a single static PV measurement – using the super syringe and the constant flow method – on gas exchange.

Materials and methods

The study was approved by our ethics committee. Seventeen consecutive patients were investigated early in the course of ARDS (<24 hour) once written informed consent had been obtained from each patient's next of kin. Patients met the following criteria: arterial oxygen tension (Pao2)/fractional inspired oxygen (Fio2) ratio of 200 or less, bilateral radiographic pulmonary infiltrates, and pulmonary artery occlusion pressure of 18 mmHg or less [5]. A computed tomography scan was performed during the preceding 12 hours to classify pulmonary infiltrates as diffuse, lobar, or patchy [6]. Patients were sedated, paralyzed and ventilated under volume control ventilation (Puritan Bennett 840; Puritan Bennett, Carlsbad, CA, USA) using the following parameters throughout the study: tidal volume at 6–7 ml/kg ideal body weight, plateau pressure (Pplat) below 35 cm H2O, Fio2 at 0.8 and PEEP at 10 cmH2O.

Patients were studied during three randomly assigned and successive 1-hour periods, two of which were after the following interventions one was a control period: a PV measurement performed using a 2 l super syringe (PVSS) and a PV measurement performed using the constant flow method (PVCF). PVSS measurement was completed in 60–90 s. The patient was disconnected from the ventilator during 3 s to reach functional residual capacity. Then, 100 ml samples of oxygen were given with a 2 s pause at the end of each inflation until an airway pressure of 60 cmH2O was achieved. Finally, 100 ml samples of oxygen were aspirated with a 2 s pause at the end of each deflation until an airway pressure of 0 cmH2O was achieved. PVCF measurement was preceded by an expiratory pause of 20 s and was completed in 8 s. Ventilatory parameters were set on zero end-expiratory pressure, a respiratory rate of 3 breaths/minute and a tidal volume of 1200 ml to obtain a constant flow of 9 l/minute, thus generating a PV curve on the screen of the ventilator [2]. The maximal peak airway pressure was set at 60 cmH2O. When a cycle at low flow was obtained, parameters of the ventilator were immediately set as initially. During the control period, patients were not disconnected from the ventilator and PEEP was unchanged.

All patients had an arterial catheter placed for monitoring of systemic pressures. Blood gases were recorded each minute via a continuous arterial sensor system (Paratrend 7; Diametrics Medical, St Paul, MN, USA) [7]. The 90% response time for the sensor is 180 s or less at 37°C [8]. In humans, the bias provided by the Paratrend 7 was found to be -1.19% for Pao2 and +1.28 mmHg for Paco2 [7]. During PVSS, inspiratory and expiratory flows were measured using a pneumotachograph (Hans-Rudolf 3700; Hans-Rudolf, Kansas City, KS, USA) and a differential pressure transducer. Airway pressures were measured using another differential pressure transducer. Volume changes were obtained by integration of the flow signal recorded using the MP100 data acquisition system (Biopac Systems, Goleta, CA, USA). A static PV curve was constructed to determinate the lower inflection point (LIP) [9] and the upper inflection point (UIP) [10]. The Chord compliance of the respiratory system (Crs) was defined as the slope of the linear part of the PV curve obtained with the super syringe technique.

Variables were expressed as mean ± standard deviation. A two-way analysis of variance (ANOVA) for repeated measures was conducted to study the effects of time and PV measurement on recorded parameters. Positive or negative responders to PV measurement were patients who exhibited an increase or a decrease in Pao2/Fio2 above 20% occurring in the 5 minutes after PV measurement and persisting for at least 15 minutes. Correlations were analyzed using Pearson product correlation. The maximal increase in Pao2 after PV measurement taken using both methods was compared between patients with diffuse, lobar, or patchy ARDS using one-way ANOVA. P < 0.05 was considered statistically significant.

Results

Characteristics of the 17 patients are summarized in Table 1. The Lung Injury Score was 3.1 ± 0.4 and the intensive care unit mortality rate was 36%. Pulmonary infiltrates were classified as diffuse in 11 patients, lobar in three and patchy in three. Tidal volume was 410 ± 96 ml (6.9 ± 1.0 ml/kg of ideal body weight) with a mean inspiratory:expiratory ratio of 1:1.9. All patients had stable haemodynamic parameters (mean arterial pressure [MAP] 76 ± 17 mmHg, heart rate 110 ± 17 beats/minute). Eight patients received norepinephrine (0.2 ± 0.1 μg/kg per minute).

Table 1 Characteristics of the patients

The insufflated volumes were 1824 ± 381 ml (range: 800–2000 ml) during PVSS and 1120 ± 115 ml (range: 820–1200 ml) during PVCF. The Pplat was 46 ± 9 cmH2O at the end of PVSS and was 46 ± 11 cmH2O at the end of PVCF. Peak airway pressure, Pplat, mean airway pressure and Crs (measured 5 minute after PV measurement) were not significantly modified after PVSS (36 ± 8 cmH2O, 27 ± 7 cmH2O, 17 ± 4 cmH2O and 58 ± 25 ml/cmH2O, respectively) and after PVCF (36 ± 6 cmH2O, 28 ± 7 cmH2O, 18 ± 4 cmH2O and 56 ± 29 cmH2O, respectively) as compared with baseline values (36 ± 6 cmH2O, 27 ± 7 cmH2O, 18 ± 4 cmH2O and 56 ± 26 ml/cmH2O, respectively).

ANOVA revealed that neither PV measurement nor time significantly affected Pao2 when measured each minute (P = 0.6 for PV measurement; P = 0.25 for time; P = 0.2 for interaction). Two patients (patients 7 and 13; Table 1) were positive responders to PVCF (Pao2/Fio2 ratio increased after PVCF by 102% in one patient and by 38% in the other; Figure 1). In one patient, Pao2 returned to baseline within 2 hours (PVCF was followed by control period) whereas the other remained a responder after 3 hours (PVCF was the last period in this patient). Two patients were negative responders to PVSS (Pao2/Fio2 ratio decreased by 40% in one patient and by 35% in the other; Figure 1). One patient remained a negative responder 60 minutes after PVSS measurement. Neither the Crs nor the Pplat reached during PV measurement was correlated with the maximal increase in Pao2 after PV measurement using both methods (data not shown). The maximal increase in Pao2 after PV measurement using both methods was not different between patients with diffuse, lobar, or patchy ARDS (data not shown).

Figure 1
figure 1

Evolution over time of Pao2/Fio2 ratio following PV measurements and during a control period. The PV measurements were taken using the super syringe (PVSS) and constant flow method (PVCF). Data are expressed as percentage increase or decrease in Pao2/Fio2 ratio at 5, 15, 30 and 60 minutes as compared with values before PV measurement or the control period. Dashed lines represent 20% increase and decrease as compared with before PV measurement. Fio2, fractional inspired oxygen; Pao2, partial arterial oxygen tension; PV, pressure-volume.

Eleven patients exhibited a LIP on the PV curve obtained using the super syringe (Table 1). The PEEP level was 2 cmH2O above the LIP on inclusion in one positive responder to PVCF, whereas no LIP was identified in the other positive responder. Seven patients exhibited an UIP (at a volume of 1542 ± 82 ml and a pressure of 39 ± 12 cmH2O). An UIP was present in the two patients exhibiting a negative response to PVSS.

PV measurement did not significantly affect Paco2 and MAP. One patient had a decrease in MAP by more than 10 mmHg for less than 5 minutes after PVSS and one patient after PVCF. After PVSS, Paco2 increased by more than 10 mmHg in two patients and returned to baseline values after 15 minutes. No case of barotrauma was observed on the chest radiograph performed on the day after the protocol.

Discussion

Taking the measurements necessary to construct a single PV curve without changing the PEEP level, either by super syringe or by constant flow method, does not durably affect gas exchange and haemodynamic parameters in a population of ARDS patients. Therefore, early evaluation of the impacts of changing ventilator settings or therapeutic interventions should not be influenced by any lasting effect of PV measurement. However, a very limited number of patients exhibit a sustained alteration in oxygenation following PV measurements. Therefore, if a small sample of patients or animals is studied, then a blood gas analysis should be performed before and after PV measurement before any therapeutic intervention is applied, in order to ensure that blood gas analysis is reliable.

Our study compared the two methods commonly used for PV measurement. PV measurement using the constant flow method was able to improve oxygenation over several hours in two of our 17 patients, whereas PV measurement using super syringe impaired Pao2 in two patients. This selective effect could be explained by the differences in the design of these PV curve methods. PV measurement using the super syringe consists of a significant phase of alveolar recruitment at inflation but this is followed by an active expiration and by a short disconnection from the ventilator that probably prevents any sustained recruitment. Moreover, this active expiration followed by disconnection could have resulted in a dramatic decrease in Pao2, although no such effect was observed in the present investigation. In a recent study Lee and coworkers [3] found that PV measurement using a super syringe was well tolerated in most ARDS patients but caused significant changes in pulse oximetry. However, this latter study did not evaluate for how long oxygenation may be affected by PV measurement. During PV measurement using the constant flow method, the removal of PEEP just before PV curve assessment probably contributed to preventing any significant beneficial effect on oxygenation. The improvement in oxygenation that we observed in two patients could be accounted for by the lack of disconnection from the ventilator and the lack of active expiration as compared with the super syringe procedure. Therefore, PV measurement using the constant flow method could result in significant recruitment in a limited number of ARDS patients.

In the present study, a single PV curve measurement did not affect oxygenation while maintaining a PEEP level of 10 cmH2O after PV measurement. Therefore, we cannot rule out the possibility that there is any beneficial influence of increasing PEEP level after PV measurement. Indeed, the effects of a recruitment manoeuvre were suggested to depend on the PEEP level that is applied after that recruitment manoeuvre [11, 12]. In our patients, maintaining the PEEP level unchanged after PV measurement might have contributed to an early loss of recruitment possibly achieved during the PV manoeuvre.

We performed only one PV measurement, and therefore we cannot rule out any deleterious effect of PV measurements repeated at short intervals. Indeed, repeated generation of a PV curve using the constant flow method in pigs subjected to lung lavage was recently shown to induce de-recruitment by repeated removal of PEEP [4].

The response to a potential recruitment manoeuvre might depend on the nature of the insult (pulmonary versus extrapulmonary) [13], and on the stage of lung disease (early versus late phase) [14]. Indeed, it is likely that a recruitment manoeuvre is less effective in pulmonary ARDS as well as in late ARDS (for example in patients with more consolidation or fibrosis) [12]. In our study we included mainly patients with pulmonary ARDS. This could have contributed to the lack of beneficial effect of constructing a PV curve on oxygenation. However, our patients presented with early and mainly diffuse ARDS, which should have potentiated the recruitment potentially induced by a PV manoeuvre.

Conclusion

The effects of a strategy aimed at improving oxygenation can be reliably recorded early after a single PV measurement that is not followed by a change in PEEP level. This finding is important because many clinical and experimental studies report early evaluation findings for therapeutic interventions that are initiated just after PV measurement. Even if a few patients exhibit a sustained improvement in oxygenation (>60 minutes) after PV measurement using the constant flow method, then this latter method cannot be considered a recruitment manoeuvre. We confirmed that PV curve assessment is well tolerated in ARDS patients.

Key messages

  • The generation of a single pulmonary PV curve, whether one uses the super syringe or the constant flow method, does not significantly and durably affect oxygenation and haemodynamic parameters in ARDS patients.

  • Evaluation of the effects of a strategy aiming at improving oxygenation can be reliably recorded early after PV measurement.

Abbreviations

ANOVA:

= analysis of variance

ARDS:

= acute respiratory distress syndrome

Crs:

= Chord compliance of the total respiratory system

Fio2:

= fractional inspired oxygen

LIP:

= lower inflection point

MAP:

= mean arterial pressure

Paco2:

= arterial partial carbon dioxide tension

Pao2:

= arterial partial oxygen tension

PEEP:

= positive end-expiratory pressure

Pplat:

= airway plateau pressure

PV:

= pressure-volume

UIP:

= upper inflection point.

References

  1. Matamis D, Lemaire F, Harf A, Brun-Buisson C, Ansquer JC, Atlan G: Total respiratory pressure-volume curves in the adult respiratory distress syndrome. Chest 1984, 86: 58-66.

    Article  CAS  PubMed  Google Scholar 

  2. Lu Q, Vieira SR, Richecoeur J, Puybasset L, Kalfon P, Coriat P, Rouby JJ: A simple automated method for measuring pressure-volume curves during mechanical ventilation. Am J Respir Crit Care Med 1999, 159: 275-282.

    Article  CAS  PubMed  Google Scholar 

  3. Lee WL, Stewart TE, MacDonald R, Lapinsky S, Banayan D, Hallett D, Mehta S: Safety of pressure-volume curve measurement in acute lung injury and ARDS using a syringe technique. Chest 2002, 121: 1595-1601. 10.1378/chest.121.5.1595

    Article  PubMed  Google Scholar 

  4. Henzler D, Mahnken A, Dembinski R, Waskowiak B, Rossaint R, Kuhlen R: Repeated generation of the pulmonary pressure-volume curve may lead to derecruitment in experimental lung injury. Intensive Care Med 2005, 31: 302-310. 10.1007/s00134-004-2512-1

    Article  PubMed  Google Scholar 

  5. Bernard GR, Artigas A, Brigham AL, Carlet J, Falke K, Hudson L, Lamy M, LeGall JR, Mois A, Spragg R: American-European consensus conference on ARDS. Am J Respir Crit Care Med 1994, 149: 818-824.

    Article  CAS  PubMed  Google Scholar 

  6. Rouby JJ, Puybasset L, Cluzel P, Richecoeur J, Lu Q, Grenier P: Regional distribution of gas and tissue in acute respiratory distress syndrome. II. Physiological correlations and definition of an ARDS Severity Score. CT Scan ARDS Study Group. Intensive Care Med 2000, 26: 1046-1056. 10.1007/s001340051317

    Article  CAS  PubMed  Google Scholar 

  7. Abraham E, Gallagher TJ, Fink S: Clinical evaluation of a multiparameter intra-arterial blood-gas sensor. Intensive Care Med 1996, 22: 507-513.

    Article  CAS  PubMed  Google Scholar 

  8. Biomedical Sensors: Paratrend 7 Operating Instructions. Malvern, PA: Biomedical Sensors; 1993:6-18.

    Google Scholar 

  9. Gattinoni L, Pesenti A, Avalli L, Rossi F, Bombino M: Pressure-volume curve of total respiratory system in acute respiratory failure. Computed tomographic scan study. Am Rev Respir Dis 1987, 136: 730-736.

    Article  CAS  PubMed  Google Scholar 

  10. Roupie E, Dambrosio M, Servillo G, Mentec H, el Atrous S, Beydon L, Bun-Buisson C, Lemaire F, Brochard L: Titration of tidal volume and induced hypercapnia in acute respiratory distress syndrome. Am J Respir Crit Care Med 1995, 152: 121-128.

    Article  CAS  PubMed  Google Scholar 

  11. Lim S, Adams AB, Simonson DA, Dries DJ, Broccard AF, Hotchkiss JR, Marini JJ: Intercomparison of recruitment maneuver efficacy in three models of acute lung injury. Crit Care Med 2004, 32: 2371-2377. 10.1097/01.CCM.0000147445.73344.3A

    Article  PubMed  Google Scholar 

  12. Kloot TE, Blanch L, Melynne Youngblood A, Weinert C, Adams AB, Marini JJ, Shapiro RS, Nahum A: Recruitment maneuvers in three experimental models of acute lung injury. Effect on lung volume and gas exchange. Am J Respir Crit Care Med 2000, 161: 1485-1494.

    Article  CAS  PubMed  Google Scholar 

  13. Pelosi P, Cadringher P, Bottino N, Panigada M, Carrieri F, Riva E, Lissoni A, Gattinoni L: Sigh in acute respiratory distress syndrome. Am J Respir Crit Care Med 1999, 159: 872-880.

    Article  CAS  PubMed  Google Scholar 

  14. Grasso S, Mascia L, Del Turco M, Malacarne P, Giunta F, Brochard L, Slutsky AS, Ranieri M: Effects of recruiting maneuvers in patients with acute respiratory distress syndrome ventilated with protective ventilatory strategy. Anesthesiology 2002, 96: 795-802. 10.1097/00000542-200204000-00005

    Article  PubMed  Google Scholar 

Download references

Acknowledgements

A grant (PHRC 2002) was obtained from the French Ministry of Health for the present work.

Author information

Authors and Affiliations

Authors

Corresponding author

Correspondence to Antoine Roch.

Additional information

Competing interests

The authors declare that they have no competing interests.

Authors' contributions

AR and LP designed the study and drafted the manuscript. AR performed the statistical analysis. AR, JMF, DD, JMA, SD and MG performed the study. All authors read and approved the final manuscript.

Authors’ original submitted files for images

Below are the links to the authors’ original submitted files for images.

Authors’ original file for figure 1

Rights and permissions

Reprints and permissions

About this article

Cite this article

Roch, A., Forel, JM., Demory, D. et al. Generation of a single pulmonary pressure-volume curve does not durably affect oxygenation in patients with acute respiratory distress syndrome. Crit Care 10, R85 (2006). https://doi.org/10.1186/cc4936